Modbus Function Specification
Scope
Applies to AirGradient ONE Modbus (8PSLM) and AirGradient ONE Display Edition with Modbus (8PSLM-DE). Note that Modbus is not supported on the Standard model.
Applies to firmware version 9.2.6 onwards.
General
  • Protocol:
    Modbus RTU / RS-485
  • Number of addresses: 1 .. 247
    Factory default: 1
  • Transmission format: 8 bits, none / odd / even parity, 1 or 2 stop bits
    Factory default; 8-E-1
  • Baudrate: 1200 .. 256000
    Factory default: 19200
Timings
Scan Rate
Depending on settings the monitor will perform a measurement cycle every few minutes and provide the newly measured values on the Modbus interface once the cycle has successfully completed. When scanning at a rate of e.g. 10s then the maximum resulting latency is 10s.
Response Timeout
The monitors typically respond within 50 .. 60ms, hence a response timeout setting of 100ms is recommended. For baudrates < 9600 better results might be achieved with a higher setting.
Delay between Polls
A setting of > 50ms between slave polls is recommended.
Pinout
The connector pinout is
Voltage
The 12V .. 32V range is supported.
Function 04 Read Registers
These are the registers that the Master can read.
Register
Parameter
Data Type
Range
Unit
Comment
3x0000
current CO2
signed 16 bit
-1 .. 10,000
ppm
raw value
-1 indicates an error condition
3x0001
current PM2.5
signed 16 bit
-1 .. 1,000
ppm
raw value
-1 indicates an error condition OR the sensor is not ready yet
3x0002
current PM10
signed 16 bit
-1 .. 1,000
ppm
raw value
3x0003
current temperature
signed 16 bit
-10001
-400 .. 900
tenth of °C, eg. 126 for 12.6°C
raw value
-10001 indicates an error condition
3x0004
current relative humidity
signed 16 bit
-1 .. 100
raw value
3x0005
current TVOC
signed 16 bit
-1 .. 30,000
OR
-1 .. 500
ppb (SGP 30)
OR
SGP 40 index, see https://sensirion.com/products/catalog/SGP40/
temperature and humidity compensated
-1 indicates an error condition, in case of SGP 30 ppb’s are delivered, in case of the SGP 40 an index
3x0010
sensorId byte #1
16 bit
0x00 .. 0xff
MSB
3x0011
sensorId byte #2
16 bit
0x00 .. 0xff
3x0012
sensorId byte #3
16 bit
0x00 .. 0xff
3x0013
sensorId byte #4
16 bit
0x00 .. 0xff
3x0014
sensorId byte #5
16 bit
0x00 .. 0xff
3x0015
sensorId byte #6
16 bit
0x00 .. 0xff
LSB
3x0020
friendly name char #1
8 bit char
ASCII
a user friendly name, up to 24 chars
3x0038
friendly name char #24
Function 05 Write Single Coil
These are the single bit registers the Master can write.
Register
Parameter
Data Type
Range
Unit
Comment
Register
Parameter
Data Type
Range
Unit
Comment
0x0000
calibrate CO2 to 400ppm
bit 0
1 / 0
the calibaration is performed upon next measurement cycle, normally within a few minutes
0x0000
led test mode
bit 1
1 / 0
not implemented yet
0x0000
Function 16 Write Multiple Registers
These are the registers the Master can write.
Register
Parameter
Data Type
Range
Unit
Comment
0x0000
8 bit char
ASCII
0x0018
